Segmented Ion Chambers

These ion chambers will consist of four copper pads on a circuit board with a common high voltage plate. These pads will be side-by-side, allowing for four simultaneous, independent measurements. There will be a 1 inch chamber for I0 and a 2 inch chamber for It.

Initial experiment

Use a foil wide enough to be measured by each channel. The goal is to make four simultaneous measurements, each of which is of the same quality as a measure of the foil using a normal set of ion chambers.

We only have 6 v2f channels at X23A2 at this time. This means:

  1. We can only test three channels at a time (no feedback, no Ir)

  2. We will need to feed back off one of the I0 channels

Other thoughts:

  • Ultimately, we will need a total of 2N+2 signal chains -- one for feedback, one for the reference, N I0 channels, and N It channels. For N=4, that means 10 signal chains. Oh boy! I get to buy stuff for the beamline!

  • We probably want vertical knife edges (or at least lead tape) defining the horizontal extent of the beam entering each I0 channel. Hutch slits will be used to define the vertical extent of the beam.

  • We will need a good way of aligning the detectors with the beam coming off the mirror. Perhaps we should start with Pd or some such...

  • I will need to write an Athena filter to load these data all at once.

Tests

  • Cross talk. Block all but one channel with lead tape and measure all four channels. The goal is to see one good measurement and three null measurements.

  • Linearity. Normal linearity test -- attenuate beam before I0, demonstrate that the ratios of each channel are constant.

  • Find an appropriate voltage for linear response

  • Map glitch behavior per channel. Do we see a glitch migrating from channel to channel as the energy changes?

  • How does the data behave when normalized just by the correct I0 channel compared to a normalization by the sum of I0 channels? That is, do we need a segmented I0 chamber? (I suspect yes, but we need to show this.)

Second experiment

Simon suggests a reactor with space for four samples meeting the geometry of the ion chambers. Experiment 2A will be to fill each sample slot with a materials and run the reactor. All four should measure the same thing. Experiment 2B is to do four independent experiments in the reactor and measure 4 things. Then we publish.

Far future

  • Fluorescence experiments are possible by staggering the samples along the beam path and pointing a fluo detector at each sample. Shielding is probably required.
  • At NSLS-II, the unfocused beam in the hutch will be 55 mm wide. Once can imagine 10 to 15 side-by-side experiments. Wow!